Compounding Errors: Combining Accuracies of Fluke Meter & Clamp

  • Thread starter Thread starter elimenohpee182
  • Start date Start date
AI Thread Summary
Compounding errors when measuring electrical current with a Fluke meter and current clamp involves understanding the accuracy specifications of both devices. The Fluke meter has an accuracy of ±0.5% + 5 counts, while the clamp has ±1% + 5mA, leading to a combined uncertainty that can be complex to calculate. The discussion highlights that the meter's resolution can significantly impact the overall accuracy, especially at lower current measurements. To improve measurement precision, using a higher resolution voltmeter, switching to a current transformer clamp, or incorporating a voltage amplifier is recommended. Accurate results depend on careful consideration of both devices' specifications and potential calibration errors.
elimenohpee182
Messages
4
Reaction score
0
Hey Guys,
I'm not sure the correct procedure for comounding errors using some test equipment. I'm trying to measure electrical current using a fluke meter and a fluke current clamp. Here are the two devices I'm using specifically:

Fluke Power Analyzer Model 437 Series 2
Fluke i5sPQ3 AC Current Clamp

Here is where it is confusing. The Fluke meter has an accuracy spec as well as the current clamp. For instance, the meter says in the range 0.5 - 200A, the accuracy is 0.5% + 5 counts. The clamp says the accuracy is 1% + 5mA in the range of 10mA - 1A.

What is the appropriate way to combine these two accuracies to establish a single accuracy? I would think to add them, but I've seen some people that seem to take an average, using a RMS-type formula.

Any suggestions?
 
Engineering news on Phys.org
Meter ±0.5% ±5 counts.
Clamp ±1% ±5mA, in the range of 10mA to 1A.

Combined ±1.5%, ±5 counts, ±5mA, over the range of 10mA to 1A.
 
Say you have 1.0 A on your meter. The meter accuracy says that the actual current measured is within the largest range between ±0.5% (0.995-1.005 A) and ±5 counts (0.5-1.5 A); 0.5-1.5 A it is.

Then the current sent by the clamp (which is anything between 0.5 A and 1.5 A) could have been within the largest range between ±1% (0.495-1.515 A) and ±5mA (0.495-1.505 A); Hence the actual current of your 1.0 A measurement could be anywhere between 0.495-1.515 A.

Doing the same procedure with a 0.5 A measurement yields:

meter:
  • ±0.5%: 0.4975-0.5025 A
  • ±5 counts: 0.0-1.0 A (largest)
clamp:
  • ±1%: 0.0-1.01 A
  • ±5mA: (-0.005)-1.005 A
Possible actual current measured: (-0.005) - 1.01 A

If you used your clamp with the 1-5 A range with a 1% accuracy, then with a 5.0 A measurement, the actual current range is:

meter:
  • ±0.5%: 4.975-5.025 A
  • ±5 counts: 4.5-5.5 A (largest)
clamp:
  • ±1%: 4.455-5.555 A
Basically, the ±5 counts on the 0.1 A resolution of your meter is killing the accuracy of your clamp.
 
elimenohpee182 said:
... the meter says in the range 0.5 - 200A, the accuracy is 0.5% + 5 counts. The clamp says the accuracy is 1% + 5mA in the range of 10mA - 1A.
When using these accuracies to determine an uncertainty of a measurement, only comparison with a standard will provide a way of eliminate or compensate for any calibration errors.
 
I note that the “Fluke i5sPQ3” AC Current Clamp appears not to be a current ratio transformer, but is specified to produce a voltage output of 400 mV/A. I expect that requires the “Fluke Model 437 Series 2” Power Analyzer then be used as a volt meter on the range up to 5A * 0.4V = 2V FSD.

But the 437-II meter range is specified as Vrms 1V to 1000V phase to neutral, with resolution 0.01 V ± 0.1% of nominal voltage.
for 0.5A * 0.4V/A = 0.2V, 0.01V is 20:1 = 5%
for 1.0A * 0.4V/A = 0.4V, 0.01V is 40:1 = 2.5%
for 2.0A * 0.4V/A = 0.8V, 0.01V is 80:1 = 1.25%
for 5.0A * 0.4V/A = 2.0V, 0.01V is 200:1 = 0.5%
So the limitation is not always the 1% accuracy of the clamp, but is the resolution of the voltmeter.

To get more accurate results it would seem necessary to use a higher resolution voltmeter, change to a current transformer clamp, or use a voltage amplifier between the clamp and analyser.
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...