Isolating an oscilloscope is crucial for safety when measuring powerline voltages, as standard oscilloscopes connect their metal parts to chassis ground, posing a shock hazard. To safely work with high voltages, an isolation transformer or a battery-powered oscilloscope is necessary to prevent dangerous potential differences with grounded objects. An isolation transformer eliminates bonding with the ground, containing shock hazards within the device itself. Disconnecting the Earth connection alone is insufficient for safety. Proper isolation techniques ensure safe measurements and protect both the user and the equipment.
elecengr
Messages
3
Reaction score
0
Hai all,
Could you please explain the need for isolating the oscilloscope. If not isolated what will happen?
 
Engineering news on Phys.org
elecengr said:
Hai all,
Could you please explain the need for isolating the oscilloscope. If not isolated what will happen?

A regular oscilloscope will have a 3-prong plug (at least in the US), and all of the external metal on the 'scope, including the ground shield and clip of the probes will be connected to chassis ground. So you cannot connect the ground clip of your 'scope to Neutral or Hot when working with powerline measurements.

To work with powerline voltages, you either need to use an isolation transformer on the mains power, or use a battery-powered (floating) oscilloscope.

Does that address your question? What is the context of your question?
 
berkeman said:
A regular oscilloscope will have a 3-prong plug (at least in the US), and all of the external metal on the 'scope, including the ground shield and clip of the probes will be connected to chassis ground. So you cannot connect the ground clip of your 'scope to Neutral or Hot when working with powerline measurements.

To work with powerline voltages, you either need to use an isolation transformer on the mains power, or use a battery-powered (floating) oscilloscope.

Does that address your question? What is the context of your question?


Thanks. I can understand. But still I've some doubts. Is it enough to disconnect Earth connection? If isolation transformer is to be used, how it protects the device?
 
elecengr said:
If isolation transformer is to be used, how it protects the device?

In electronics testing, troubleshooting and servicing, an isolation transformer is a 1:1 power transformer which is used for safety. Grounded objects (desk, lamp, concrete floor, oscilloscope ground lead, etc.) near a device under test which is not isolated may be at a hazardous potential difference with respect to that device. By using an isolation transformer, the bonding is eliminated, and the shock hazard is entirely contained within the device—there is no danger in touching a live part of the circuit while another part of the body is at Earth potential.

http://en.wikipedia.org/wiki/Isolation_transformer"
 
Last edited by a moderator:
Hopefully you wouldn't need to float the whole oscilloscope.

Dual trace oscilloscopes have two vertical inputs which you can usually invert or add with front panel switches.

So, if you want to measure the voltage drop across a resistor in a mains circuit or anywhere else (and assuming you do this with great care) you put one probe on one side of the resistor and one on the other. Then invert one of the inputs and add the two inputs.

This is the same as subtracting one of the inputs from the other.
 
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
Thread 'Beauty of old electrical and measuring things, etc.'
Even as a kid, I saw beauty in old devices. That made me want to understand how they worked. I had lots of old things that I keep and now reviving. Old things need to work to see the beauty. Here's what I've done so far. Two views of the gadgets shelves and my small work space: Here's a close up look at the meters, gauges and other measuring things: This is what I think of as surface-mount electrical components and wiring. The components are very old and shows how...

Similar threads

Back
Top