A What exactly is the detector of an ellipsometer measuring?

  • A
  • Thread starter Thread starter Gabriel Maia
  • Start date Start date
  • Tags Tags
    Detector Measuring
AI Thread Summary
An ellipsometer measures the polarization state of light after it reflects off a sample, which can be a surface, interface, or thin film. The detector primarily measures the intensity of the light, capturing how it varies as the polarizer and analyzer orientations change. This modulation of intensity provides critical information about the sample's properties. Measurements typically involve multiple angles, including the angle between the polarizer and analyzer, as well as the incidence angle on the sample. Thus, the detector's readings are not based on a single measurement but rather on a series of intensity variations.
Gabriel Maia
Messages
70
Reaction score
1
Hi. I'm curious about how a PSA (polariser-sample-analyser) ellipsometer works, especifically the detector bit. In an ellipsomenter, light passes through a polariser which puts it into a known polarisation state. After that it hits a sample (which may be an interface, a surface or a thin film) and because the Fresenel coefficients behave differently for different polarisation states, the components of the light interacting with the sample are modified differently, making the reflected light elliptically polarised. Light is then passed through an analyser and is collected by the detector.

After the analyser, the electric field has an amplitude given by the sum of the components of the electric field before the analyser and is oscillating in the direction prescribdeb by the analyser.

Now, my doubt is... what is the detector really measuring? Is it measuring the light intensity? How a single measurement can give any information about the sample?

Thank you very much.
 
Physics news on Phys.org
Simply optical power. There are various choices as to how to measure optical power, but for a single beam system this can just be a photodiode. For imaging systems it can be a camera. All of the information is in the modulation of the intensity as a function of polarizer - analyzer orientation.
 
Sorry for the double post. I didn't see the "single measurement" question. It isn't a single measurement. The polarizer or analyzer spins and the intensity is measured as a function of angle. In some systems a waveplate is spun. In some systems the variable waveplate is electro optical for fast measurement. In some systems a single orientation of analyzer and compensator are found that null the signal. While this is only one position it still takes multiple measurements to find it.
 
Cutter Ketch said:
Sorry for the double post. I didn't see the "single measurement" question. It isn't a single measurement. The polarizer or analyzer spins and the intensity is measured as a function of angle. In some systems a waveplate is spun. In some systems the variable waveplate is electro optical for fast measurement. In some systems a single orientation of analyzer and compensator are found that null the signal. While this is only one position it still takes multiple measurements to find it.
Thank you very much. The angle the intensity is a function of is the angle between polariser and analyser, right? Don't they do measurements based on the incidence angle on the sample as well?
 
Thread 'Is there a white hole inside every black hole?'
This is what I am thinking. How much feasible is it? There is a white hole inside every black hole The white hole spits mass/energy out continuously The mass/energy that is spit out of a white hole drops back into it eventually. This is because of extreme space time curvature around the white hole Ironically this extreme space time curvature of the space around a white hole is caused by the huge mass/energy packed in the white hole Because of continuously spitting mass/energy which keeps...