Can pulse generators output 100mA?

AI Thread Summary
The output current of a pulse generator depends on its specifications, which are often not detailed in datasheets. A 5V pulse with a 50-ohm output impedance theoretically allows for 100mA output, but this is contingent on the pulse duration being sufficiently long. If the load is 1 ohm, one could expect around 100mA, assuming the pulse generator is designed for that output. To verify the output impedance, one should measure the open circuit pulse amplitude and then apply a 50-ohm load, observing the voltage drop. Understanding the specific model and manufacturer of the pulse generator is crucial for accurate assessment.
Bakez
Messages
11
Reaction score
0
If I have a pulse generator what is the current it can output? It doesn't say in the datasheets

IF I have a 5V pulse and 50ohm output impedence am I supposed to believe it outputs 100mA?

And then if my load is 1ohm I should get something near 100mA through it?
 
Engineering news on Phys.org
If I have a pulse generator what is the current it can output? It doesn't say in the datasheets
That depends on your pulse generator.

IF I have a 5V pulse and 50ohm output impedence am I supposed to believe it outputs 100mA?

And then if my load is 1ohm I should get something near 100mA through it?
If the pulse is not too short, I would expect this to work.
 
Who is the manufacturer and what is the model number of your pulse generator ?
 
100ns pulse
 
Nobody can answer that question without a data sheet. If this is an instrument, then it may have a 50 ohm output impedance, so analyze as if there were a 50 ohm resistor in series. Or it may have settings for the output impedance, or a low impedance mode. I expect 50 ohms.

To test it, measure the open circuit pulse amplitude, then apply a 50 ohm load. The change in voltage will tell you the output impedance. It will go to 1/2 the voltage if it is a 50 ohm generator.
 
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
suppose you have two capacitors with a 0.1 Farad value and 12 VDC rating. label these as A and B. label the terminals of each as 1 and 2. you also have a voltmeter with a 40 volt linear range for DC. you also have a 9 volt DC power supply fed by mains. you charge each capacitor to 9 volts with terminal 1 being - (negative) and terminal 2 being + (positive). you connect the voltmeter to terminal A2 and to terminal B1. does it read any voltage? can - of one capacitor discharge + of the...
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...
Back
Top