How altitude effects gas flow rate or not?

AI Thread Summary
The discussion focuses on how altitude affects gas flow rates, specifically comparing conditions at sea level (1013 mBar) and at 10,000 feet (700 mBar) with a constant oxygen source. The setup involves a variable flow restrictor initially set for a volumetric flow rate of 10 L/min at sea level, which remains unchanged at altitude. It is suggested that if the flow is choked, downstream pressure should not affect the flow rate, but experimental results indicate otherwise, showing an increase in flow proportional to absolute pressure ratios up to 20,000 feet. The geometry of the flow restriction, described as an annulus with a tapered shaft, raises questions about whether it behaves like a thin plate orifice and how converging or diverging sections impact choked flow. Overall, the relationship between altitude, pressure, and flow characteristics remains complex and warrants further exploration.
jango
Messages
2
Reaction score
0
I am considering the effects on volumetric and mass flow rate for two conditions:

Condition 1) 1013mBar, 21C (Air density as per)
Condition 2) 700mBar (10,000ft) 21C (Air density as per)

Setup: A variable flow restrictor is connected to a constant source of oxygen at 50 psig. The variable flow restrictor is adjusted to give a volumetric flow rate of 10 L/min for condition 1.

Question: The same setup is now in condition 2. What happens to the flowrate?

Note 1: The variable restrictor is not adjusted when at condition 2. It remains as it was set for condition 1.
Note 2: It appears that the flow is 'choked' as the difference between upstream pressure and downstream pressure stays within the limits required to be in the choked state (Ratio1.9±0.2)

I'm unsure how the density ratio for the two conditions could affect the flow output?
 
Engineering news on Phys.org
If the flow is choked (and that certainly appears to be), the downstream pressure will not affect the flow rate.
 
What cjl says is certainly true, though it will depend on the geometry of your exit. If it is simply a straight section or a converging section then your flow in both situations is choked (Mach 1 at the region of smallest area) and therefore lowering the downstream pressure is not going to change the mass flux. Only changing the area of your exit or your upstream pressure would change your mass flux.

If your exit is diverging, you would have additional work to do since at least a portion of the flow in the nozzle will be supersonic and the downstream pressure required to choke the flow will change.
 
Thanks cjl & boneh3ad.

I thought the restriction was choked, but experimental results suggest otherwise.

The geometry is as follows:

The restriction is created by an annulus made up by a tapered shaft in a plain hole.
The hole is 1.58mm in diameter.
The plate thickness is 0.85mm.
The shaft within the hole tapers from approx diameter 1.50mm to 1.49mm within the o.85mm thickness.

It would appear that the flow increases directly proportional to the absolute pressure ratio up to altitudes of 20,000ft (465mBar).

This surprised me because I was under the impression that as long as the abs pressure ratio was above about 2, the downstream pressure would not change the flow.

Can you explain a little more about the converging / diverging effects on choked flow? From the restriction description, is this considered a thin plate orifice?

Thanks in advance.
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
Thread 'How can I find the cleanout for my building drain?'
I am a long distance truck driver, but I recently completed a plumbing program with Stratford Career Institute. In the chapter of my textbook Repairing DWV Systems, the author says that if there is a clog in the building drain, one can clear out the clog by using a snake augur or maybe some other type of tool into the cleanout for the building drain. The author said that the cleanout for the building drain is usually near the stack. I live in a duplex townhouse. Just out of curiosity, I...