Differential Pressure Transmitter (DPT) Calibration

AI Thread Summary
To calibrate a 4-20mA Differential Pressure Transmitter (DPT) for a range of 5 to 20 psi, a calibrated pressure pump is essential. The calibration process involves teaching the instrument that 5 psi differential pressure corresponds to 4 mA and 20 psi corresponds to 20 mA. For smart transmitters, a HART communicator can be utilized to set the Upper Range Value (URV) and Lower Range Value (LRV) as needed. Additional literature and resources are available for further guidance on calibration procedures. Proper calibration ensures accurate pressure measurements in various applications.
JohnielWhite
Messages
44
Reaction score
0
I understand the basic principle behind pressure sensor calibration but i am still not able to do a calibration problem. Could someone Explain how you would go about calibrating a standard 4-20mA Differential Pressure Transmitter (DPT) to measure differential pressure in the range 5 to 20 psi?
Or direct me to literature that may be helpful in solving this problem. Thanks in Advance.
 
Physics news on Phys.org
A calibrated pressure pump is often used. With the calibrated pressure pump, you would "teach" the instrument that 5 psid is 4 mA and 20 psid is 20 mA.
 
thanks for replying...
 
if you are using a smart transmitter you can use a hart communicator ..set URV and LRV according to your requirement..

pls refer the following link for more information..
http://instrumentpedia.blogspot.in/2012/04/calibration-procedure-differential.html
 