Probing high speed digital circuits
To avoid reflections when probing a 100MHz bus from CPU to Memory, it's crucial to use a high-impedance adapter with the logic analyzer. The Tek TLA5204B and P6410 probe are suitable for this purpose, as the probe's specifications indicate a low loading effect of 2pF and 20kOhms. Keeping connections short between the bus and the probe is essential to minimize signal degradation. The absence of termination on the bus necessitates careful probing techniques to ensure signal integrity. Proper setup and equipment selection will help maintain accurate signal measurements without interference.
likephysics
I have a 100MHz bus going from CPU to Memory. The routing length is about 2inches. So there is no termination.
When I hook up the bus to a Logic analyzer, how do I know the Logic analyzer probes are not causing any reflection and messing up the bus signals. The length of the probe is about 40inches(1meter).
The probes are connected to wires(~4inches) that are soldered to the bus traces on the PCB.
What would be the right way to probe these signals.
The logic analyzer probe head should have a high-impedance adapter thing. What logic analyzer are you using? What adapter modules are available for it? For the old HP 16500 logic analyzers that we have in the lab, the adapter modules are 01650-63201 (or -63203):
http://www.nationaltestequipment.com/unitDetails.cfm?modelID=8805
We are using the tek TLA5204B
The probe is P6410
http://www.tek.com/products/logic_analyzers/tla5000/
The P6410 manual at the Tek website shows the equivalent circuit for the probe points as 2pF // 20kOhms. That should not load your bus signals too much. Be sure to keep the connections short between the bus and the probe.
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
suppose you have two capacitors with a 0.1 Farad value and 12 VDC rating. label these as A and B. label the terminals of each as 1 and 2. you also have a voltmeter with a 40 volt linear range for DC. you also have a 9 volt DC power supply fed by mains. you charge each capacitor to 9 volts with terminal 1 being - (negative) and terminal 2 being + (positive). you connect the voltmeter to terminal A2 and to terminal B1. does it read any voltage? can - of one capacitor discharge + of the...
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...